Dear Larry and Chandralal,
This is a very interesting topic in vaccine storage temperature monitoring at stores and facilities.
From my understanding, temperature alarms in a 30-day temperature loggers indicate a deviation of storage temperature outside the recommended range of +2 to +8 centigrade.
Based on the 4 high temperature alarms you observed, a health workers should understand that a 30-day device has recorded the storage temperature in a CCE went 4-times out of the recommended range. This means that, there was four incidences wherer vaccines were expoused to high temperatures (above +8 Centigrade).
According to the manufacturer, the high temperature alarm triggers when the CCE records an exposure of vaccine to +8 Centigrade for duration of +10 hours consecutively. The low temperature alarm is triggered when the temperature is -0.5 Centigrade for +60 minutes consecutively. However, I personally had observation of some devices which recorded an alarm when the exposure duration has not reached +10 hours!!!
From my understanding, the health worker doesn't need to throw away the vaccine unless the Vaccine Vial Monitor(s) (VVM) sticker on a vaccine vial(s) has reached the discard point. In addition, the alarm should equally be taken into consideration for interpretetion to determine issues like duration of exposures, times of exposure etc.
If there a frequency of repeated exposures, a health worker should have a plan to move vaccines to other facilities so as to minimize or avoid wastages.
Regards,
Bonaventura